Because of the high maintenance needed to monitor and filter spammers from the RF Cafe Forums, I decided that it would be best to just archive the pages to make all the good information posted in the past available for review. It is unfortunate that the scumbags of the world ruin an otherwise useful venue for people wanting to exchanged useful ideas and views. It seems that the more formal social media like Facebook pretty much dominate this kind of venue anymore anyway, so if you would like to post something on RF Cafe's Facebook page, please do.

Below are all of the forum threads, including all the responses to the original posts.


 Post subject: Group Delay Testing Issues
Posted: Wed Apr 19, 2006 1:35 pm 
 
Lieutenant

Joined: Wed Apr 19, 2006 1:27 pm
Posts: 3
I'm testing a fiber optic TX/RX pair with a long (~5km) fiber link. I'm using the 8510C to get the group delay, but it's giving me a negative number. When I test the same TX/RX pair with a short (1m) fiber, I get a 15ns delay. With the 5km fiber, the delay is ~-80ns. I don't know why I'm getting this result . . . any advice would be appreciated.

Thanks,

Dan


 
   
 
 Post subject: Group Delay
Posted: Wed Apr 19, 2006 4:10 pm 
 
Captain

Joined: Fri Feb 17, 2006 6:25 pm
Posts: 11
Location: US
Dan,

Your problem is due to insufficient sampling points in the measurement. Group delay is measured as the slope of the phase vs frequency. If there are too few sampled points, the instrument is confused. This occurs because a typical phase plot slopes down to -180 degrees and then restarts at +180 degrees. Without several points within each 360 degree cycle the phase plot will sometimes slope upward giving a negative delay. Also, long delays have a very high phase slope, changing rapidly vs frequency.

To illustrate: a typical phase plot will have many points near the 180 degree crossing. ( -175,-176,-177,-178,-179,-180,+179,+178...) a phase plot with too few samples may go -120,+140,-10... It becomes difficult to get the correct slope because of the missed crossings at -180/+180 degrees. The worst case may be 1 point or less in each 360 degree cycle. In that case the delay is no longer a usefull measurement. The result is a plot with incorrect slope resulting in smaller than actual delay or even negative delay.

To get a better reading, increase the number of frequency sampling points and and reduce the bandwidth. For long delays it is often necessary to measure in narrow bandwidths


 
   
 
 Post subject: Group Delay
Posted: Mon Apr 24, 2006 11:14 am 
 
Lieutenant

Joined: Wed Apr 19, 2006 1:27 pm
Posts: 3
Joe,

Thanks for your advice. I've done what you suggested and now get a positive measurement. There's still an issue, though- the signal is very noisy (+/-10ns), and I'd like to get that down to 1ns to see the delay ripple across any given band. Is this possible with the 8510C? I've done grop delay using spec ans and not had this kind of issue . . .

Thanks,

Dan


 
   
 
 Post subject: Group Delay
Posted: Tue Apr 25, 2006 1:55 pm 
 
Captain

Joined: Fri Feb 17, 2006 6:25 pm
Posts: 11
Location: US
Dan,

You didn't mention how much loss you had. If there is a lot of loss and the output level is low, it will be noisy. You could increase the source power (watch for saturation or compression of the unit you are measuring).

You could also use averaging to clean up the signal. This will average several sweeps and limit the effects of random noise. Don't use smoothing because this distorts the actual measurement.

Joe


 
   
 
 Post subject: Group Delay
Posted: Tue Apr 25, 2006 2:31 pm 
 
Lieutenant

Joined: Wed Apr 19, 2006 1:27 pm
Posts: 3
Joe,

Power is actually relatively high- near compression. I read elsewhere that, due to the length of the line, I won't be able to use an 8510C to measure within +/- 1 ns because I have to reduce the BW so much that it affects the measurement accuracy. This was supported in older HP documentation, claiming that I wouldn't be able to get beyond 10ns accuracy for the smaller (kHz) BWs. So, it looks like I'll have to live with what I have. Thanks for your help!

- Dan




Posted  11/12/2012
A Disruptive Web Presence

Custom Search
Over 10,000 pages indexed! (none duped or pirated)

Read About RF Cafe
Webmaster: Kirt Blattenberger
    KB3UON

RF Cafe Software

RF Cascade Workbook
RF Cascade Workbook is a very extensive system cascaded component Excel workbook that includes the standard Gain, NF, IP2, IP3, Psat calculations, input & output VSWR, noise BW, min/max tolerance, DC power cauculations, graphing of all RF parameters, and has a graphical block diagram tool. An extensive User's Guide is also included. - Only $35.
RF system analysis including
frequency conversion & filters

RF & EE Symbols Word
RF Stencils for Visio

Product & Service Directory
Personally Selected Manufacturers
RF Cafe T-Shirts & Mugs

RF Cafe Software

Calculator Workbook
RF Workbench
Smith Chart™ for Visio
Smith Chart™ for Excel
Your RF Cafe Progenitor & Webmaster