Seth Post subject: Noise figure measurements Posted: Thu Mar 19, 2009 2:56 am
Lieutenant
Joined: Sun Nov 02, 2008 10:16 pm Posts: 1 My question is on sub 1 dB noise figure measurements. Although I am using a low ENR source and an industry standard noise figure analyzer, I get the impression that the noise figure readings are consistently a little on the high side (0.2-0.3 dB higher) on the low noise amplifiers that I am measuring. Is there an accurate way to cross check the measurements that I am making? The instrument noise figure at the frequency of measurement is about 8.5 dB. Does one have to use a preamplifier to get more accurate results?
Top
nubbage Post subject: Re: Noise figure measurementsPosted: Sat Mar 21, 2009 6:32 am
General
Joined: Fri Feb 17, 2006 12:07 pm Posts: 218 Location: London UK Hi Seth I had a look at graph 6.1 of Agilent's App Note 57-2 "Noise Figure Measurement Accuracy". It estimates around 0.3dB error when using an analyzer with an 8.5dB NF. You might expect an error of around 0.2dB if you place a good 2dB NF pre-amp in front of the analyzer.
_________________ At bottom, life is all about Sucking in and blowing out.
Top
IR Post subject: Re: Noise figure measurementsPosted: Sat Mar 21, 2009 3:55 pm
Site Admin
Joined: Mon Jun 27, 2005 2:02 pm Posts: 373 Location: Germany You can do the following:
1. Compare the measurements between different DUT's and get a insight on the error magnitude. 2. Use another unit of the same Noise Figure Analyzer which you currently use and see if you get the same size of error - then most probably it is a measurement error of the Analyzer. 3. Better not to use another preamplifier, because then you add another error factors to your measurement setup (Gain and NF).
_________________ Best regards,
- IR